How do I regulate the flow in an ion exchange column?
Not enough information to answer you properly. how ever I suggest to wash (high speed) 100 cm/hr and when you have to work slowly go down to a flow of about 50-60 cm/hr. Indeed as mentioned above the best way to regulate the column flow is by a peristaltic pump. Good Luck
Not enough information to answer you properly. how ever I suggest to wash (high speed) 100 cm/hr and when you have to work slowly go down to a flow of about 50-60 cm/hr. Indeed as mentioned above the best way to regulate the column flow is by a peristaltic pump. Good Luck
Of cause a peristaltic pump with the variation of spead. But if not enogh (too high or too slow) you may change the diameter of the tubes applied in the pump. You may calibrate it by weighing the passed liquid volume collected in given times.
Of cause a peristaltic pump with the variation of spead. But if not enogh (too high or too slow) you may change the diameter of the tubes applied in the pump. You may calibrate it by weighing the passed liquid volume collected in given times.
The ideal flow rate for an ion exchange column is 1 mL.min.cm-2. You must know the cross section of the column to calculate this flow. The control must be done by a pulse or peristaltic pump.
The ideal flow rate for an ion exchange column is 1 mL.min.cm-2. You must know the cross section of the column to calculate this flow. The control must be done by a pulse or peristaltic pump.
Dear Ronny, Are you using a glass column, how is liquid passed through the column. What is your detection technique. Please provide more details. Kind regards, Ade Kujore Marketing Cecil Instruments Limited Milton Technical Centre, Cambridge CB24 6AZ United Kingdom email:- ade.kujore@cecilinstruments.com telephone:- +44 (0) 1223 420821 fax.:- +44 (0) 1223 420475 web site:- www.cecilinstruments.com
Dear Ronny, Are you using a glass column, how is liquid passed through the column. What is your detection technique. Please provide more details. Kind regards, Ade Kujore Marketing Cecil Instruments Limited Milton Technical Centre, Cambridge CB24 6AZ United Kingdom email:- ade.kujore@cecilinstruments.com telephone:- +44 (0) 1223 420821 fax.:- +44 (0) 1223 420475 web site:- www.cecilinstruments.com
Hi Ronny, It depends on the amount of sample. If you work with low volumes like me, you can easily fill a colum with your resin and adjust the effluent flow by a valve at the bottom of the column. This way you prevent flow channelling too because you have a head of water above you resin bed. Picture is attached. Hope this helps you.
Hi Ronny, It depends on the amount of sample. If you work with low volumes like me, you can easily fill a colum with your resin and adjust the effluent flow by a valve at the bottom of the column. This way you prevent flow channelling too because you have a head of water above you resin bed. Picture is attached. Hope this helps you.
Not enough information to answer you properly. how ever I suggest to wash (high speed) 100 cm/hr and when you have to work slowly go down to a flow of about 50-60 cm/hr. Indeed as mentioned above the best way to regulate the column flow is by a peristaltic pump.
Good Luck
Not enough information to answer you properly. how ever I suggest to wash (high speed) 100 cm/hr and when you have to work slowly go down to a flow of about 50-60 cm/hr. Indeed as mentioned above the best way to regulate the column flow is by a peristaltic pump.
Good Luck
More
VOTE
Of cause a peristaltic pump with the variation of spead. But if not enogh (too high or too slow) you may change the diameter of the tubes applied in the pump. You may calibrate it by weighing the passed liquid volume collected in given times.
Of cause a peristaltic pump with the variation of spead. But if not enogh (too high or too slow) you may change the diameter of the tubes applied in the pump. You may calibrate it by weighing the passed liquid volume collected in given times.
More
VOTE
By using pump you can regulate flow but if you are using present column then before experiment you have to do some trails and fix this this problem.
By using pump you can regulate flow but if you are using present column then before experiment you have to do some trails and fix this this problem.
More
VOTE
The ideal flow rate for an ion exchange column is 1 mL.min.cm-2. You must know the cross section of the column to calculate this flow. The control must be done by a pulse or peristaltic pump.
The ideal flow rate for an ion exchange column is 1 mL.min.cm-2. You must know the cross section of the column to calculate this flow. The control must be done by a pulse or peristaltic pump.
More
VOTE
Dear Ronny,
Are you using a glass column, how is liquid passed through the column. What is your detection technique. Please provide more details.
Kind regards,
Ade Kujore
Marketing
Cecil Instruments Limited
Milton Technical Centre, Cambridge CB24 6AZ
United Kingdom
email:- ade.kujore@cecilinstruments.com
telephone:- +44 (0) 1223 420821
fax.:- +44 (0) 1223 420475
web site:- www.cecilinstruments.com
Dear Ronny,
Are you using a glass column, how is liquid passed through the column. What is your detection technique. Please provide more details.
Kind regards,
Ade Kujore
Marketing
Cecil Instruments Limited
Milton Technical Centre, Cambridge CB24 6AZ
United Kingdom
email:- ade.kujore@cecilinstruments.com
telephone:- +44 (0) 1223 420821
fax.:- +44 (0) 1223 420475
web site:- www.cecilinstruments.com
More
VOTE
Peristaltic pump will be a better option
Peristaltic pump will be a better option
More
VOTE
Hi Ronny,
It depends on the amount of sample. If you work with low volumes like me, you can easily fill a colum with your resin and adjust the effluent flow by a valve at the bottom of the column. This way you prevent flow channelling too because you have a head of water above you resin bed. Picture is attached. Hope this helps you.
Hi Ronny,
It depends on the amount of sample. If you work with low volumes like me, you can easily fill a colum with your resin and adjust the effluent flow by a valve at the bottom of the column. This way you prevent flow channelling too because you have a head of water above you resin bed. Picture is attached. Hope this helps you.