Dear Andie, generally the results expressed as % ratio range between 0 to 100. For example, in the DPPH antioxidant assay 0 % inhibition means no antioxidant activity while 100 % high antioxidant activity. Maybe you got issues in the analysis of the blank sample (control).
Dear Andie, generally the results expressed as % ratio range between 0 to 100. For example, in the DPPH antioxidant assay 0 % inhibition means no antioxidant activity while 100 % high antioxidant activity. Maybe you got issues in the analysis of the blank sample (control).
It is highly unlikely to obtain negative %inhibition in DPPH assay, unless you followed a wrong protocol/formula. You may recheck the following points and revisit the assay protocol: 1. Blank for base line correction (auto zero) should be base solvent used to prepare DPPH solution. Base solvent for DPPH and sample dilution should be same/soluble. 2. Turbidity/phase separation in samples after addition of DPPH solution to reaction tubes. Check if this is the case, Asample > Acontrol which is erroneous, consider using the same solvent as described above. 3. Formula used to calculate %Inhibition (it should be [(Acontrol - Asample)/(Acontrol)]*100 Hope this may help to standardize your protocol. Wish you good luck with your experiments. Cheers... :)
It is highly unlikely to obtain negative %inhibition in DPPH assay, unless you followed a wrong protocol/formula. You may recheck the following points and revisit the assay protocol: 1. Blank for base line correction (auto zero) should be base solvent used to prepare DPPH solution. Base solvent for DPPH and sample dilution should be same/soluble. 2. Turbidity/phase separation in samples after addition of DPPH solution to reaction tubes. Check if this is the case, Asample > Acontrol which is erroneous, consider using the same solvent as described above. 3. Formula used to calculate %Inhibition (it should be [(Acontrol - Asample)/(Acontrol)]*100 Hope this may help to standardize your protocol. Wish you good luck with your experiments. Cheers... :)
Hi there, Such situation might be observed in case of activation of the oxydation by the treatment. Is there a correlation between this "activation" and dilution factor?
Hi there, Such situation might be observed in case of activation of the oxydation by the treatment. Is there a correlation between this "activation" and dilution factor?
check that the absorbance of your negative control is HIGHER than your sample, if it is not- then you can not use the formula [(Acontrol - Asample)/(Acontrol)]*100
Always prepare fresh reagents, and use them immediately. keep them closed and covered in aluminum foil (to prevent oxidization in air and light)
centrifuge your extracts before you take ur reading. turbidity can give you false results
if your raw data readings are negative- then you need to dilute your sample alot more!!!! it may be too strong, thus scavenging too quickly!!
check that the absorbance of your negative control is HIGHER than your sample, if it is not- then you can not use the formula [(Acontrol - Asample)/(Acontrol)]*100
Always prepare fresh reagents, and use them immediately. keep them closed and covered in aluminum foil (to prevent oxidization in air and light)
centrifuge your extracts before you take ur reading. turbidity can give you false results
if your raw data readings are negative- then you need to dilute your sample alot more!!!! it may be too strong, thus scavenging too quickly!!
Dear Andie, generally the results expressed as % ratio range between 0 to 100. For example, in the DPPH antioxidant assay 0 % inhibition means no antioxidant activity while 100 % high antioxidant activity. Maybe you got issues in the analysis of the blank sample (control).
Dear Andie, generally the results expressed as % ratio range between 0 to 100. For example, in the DPPH antioxidant assay 0 % inhibition means no antioxidant activity while 100 % high antioxidant activity. Maybe you got issues in the analysis of the blank sample (control).
More
VOTE
Thank you very much for your answers!
Thank you very much for your answers!
More
VOTE
What will we do if the IC50 values become negative after interepretation in excel?
What will we do if the IC50 values become negative after interepretation in excel?
More
VOTE
It is highly unlikely to obtain negative %inhibition in DPPH assay, unless you followed a wrong protocol/formula. You may recheck the following points and revisit the assay protocol:
1. Blank for base line correction (auto zero) should be base solvent used to prepare DPPH solution. Base solvent for DPPH and sample dilution should be same/soluble.
2. Turbidity/phase separation in samples after addition of DPPH solution to reaction tubes. Check if this is the case, Asample > Acontrol which is erroneous, consider using the same solvent as described above.
3. Formula used to calculate %Inhibition (it should be [(Acontrol - Asample)/(Acontrol)]*100
Hope this may help to standardize your protocol.
Wish you good luck with your experiments. Cheers... :)
It is highly unlikely to obtain negative %inhibition in DPPH assay, unless you followed a wrong protocol/formula. You may recheck the following points and revisit the assay protocol:
1. Blank for base line correction (auto zero) should be base solvent used to prepare DPPH solution. Base solvent for DPPH and sample dilution should be same/soluble.
2. Turbidity/phase separation in samples after addition of DPPH solution to reaction tubes. Check if this is the case, Asample > Acontrol which is erroneous, consider using the same solvent as described above.
3. Formula used to calculate %Inhibition (it should be [(Acontrol - Asample)/(Acontrol)]*100
Hope this may help to standardize your protocol.
Wish you good luck with your experiments. Cheers... :)
More
VOTE
Hi there,
Such situation might be observed in case of activation of the oxydation by the treatment. Is there a correlation between this "activation" and dilution factor?
Hi there,
Such situation might be observed in case of activation of the oxydation by the treatment. Is there a correlation between this "activation" and dilution factor?
More
VOTE
Revisit your protocol and always use freshly prepared reagents.
Revisit your protocol and always use freshly prepared reagents.
More
VOTE
Pls check your protocol.
Pls check your protocol.
More
VOTE
Hi Andie. Just a few tips to consider:
- check that the absorbance of your negative control is HIGHER than your sample, if it is not- then you can not use the formula [(Acontrol - Asample)/(Acontrol)]*100
- Always prepare fresh reagents, and use them immediately. keep them closed and covered in aluminum foil (to prevent oxidization in air and light)
- centrifuge your extracts before you take ur reading. turbidity can give you false results
- if your raw data readings are negative- then you need to dilute your sample alot more!!!! it may be too strong, thus scavenging too quickly!!
i hope this helps! good luck!Hi Andie. Just a few tips to consider:
- check that the absorbance of your negative control is HIGHER than your sample, if it is not- then you can not use the formula [(Acontrol - Asample)/(Acontrol)]*100
- Always prepare fresh reagents, and use them immediately. keep them closed and covered in aluminum foil (to prevent oxidization in air and light)
- centrifuge your extracts before you take ur reading. turbidity can give you false results
- if your raw data readings are negative- then you need to dilute your sample alot more!!!! it may be too strong, thus scavenging too quickly!!
i hope this helps! good luck!More