Which kind of GC column (Porapak Q or MS-5A) is the best to...
MS-5A column is definitely suitable for high-resolution separation of permanent gases but Porapak Q column usually use in food industry analyzes. So I prefer MS-5A column for separation of H2 and O2 from water splitting reaction and I would suggest GC-MS measurement instead of GC for this purpose.
MS-5A column is definitely suitable for high-resolution separation of permanent gases but Porapak Q column usually use in food industry analyzes. So I prefer MS-5A column for separation of H2 and O2 from water splitting reaction and I would suggest GC-MS measurement instead of GC for this purpose.
The problem is not to separate H2 and O2, but to separate O2 and N2. The latter is always present as a contaminant from air. The best would be a molecular sieves capillary column with Ar and TCD. The T< 100C is advised. Just try. Don't forget to bake your column from time to time at ~ 180C overnight to remove adsorbed water.
The problem is not to separate H2 and O2, but to separate O2 and N2. The latter is always present as a contaminant from air. The best would be a molecular sieves capillary column with Ar and TCD. The T< 100C is advised. Just try. Don't forget to bake your column from time to time at ~ 180C overnight to remove adsorbed water.
MS-5A column is definitely suitable for high-resolution separation of permanent gases but Porapak Q column usually use in food industry analyzes. So I prefer MS-5A column for separation of H2 and O2 from water splitting reaction and I would suggest GC-MS measurement instead of GC for this purpose.
MS-5A column is definitely suitable for high-resolution separation of permanent gases but Porapak Q column usually use in food industry analyzes. So I prefer MS-5A column for separation of H2 and O2 from water splitting reaction and I would suggest GC-MS measurement instead of GC for this purpose.
More
VOTE
The problem is not to separate H2 and O2, but to separate O2 and N2. The latter is always present as a contaminant from air. The best would be a molecular sieves capillary column with Ar and TCD. The T< 100C is advised. Just try. Don't forget to bake your column from time to time at ~ 180C overnight to remove adsorbed water.
The problem is not to separate H2 and O2, but to separate O2 and N2. The latter is always present as a contaminant from air. The best would be a molecular sieves capillary column with Ar and TCD. The T< 100C is advised. Just try. Don't forget to bake your column from time to time at ~ 180C overnight to remove adsorbed water.
More
VOTE
Sina Safavi Thank you very much for your answer. It was helpful.
Sina Safavi Thank you very much for your answer. It was helpful.
More
VOTE
Yurii V Geletii : Professor, thank you very much for your answer. I will definitely try it.
Yurii V Geletii : Professor, thank you very much for your answer. I will definitely try it.